Different CMFs with Spectros

 
 
Page  Page 2 of 2 :  « Previous  1  2

Author wikihua
ZRO

#16 | Posted: 12 Nov 2024 14:17 
Steve:
That's irrelevant, as ColourSpace doesn't (presently) use spectral data.

Hi Steve,

Thanks for clarifying. I was wondering if it might be due to the restriction to using only the XYZ data that X-Rite officially supports, rather than the full spectral data that's accessible via some open-source tools like Argyll. Thanks

Author Steve Male
INF

#17 | Posted: 14 Nov 2024 11:19 
Spectral data is not really of use for calibration - it needs to be converted into XYZ/Yxy data.
So for probes that have that as their standard output, it makes sense to use that, and no bother with spectral data.

However, we are going to add Spectral Graphs to ColourSpace, so at that point we will natively use spectral data too.

Steve
Steve Shaw
Mob Boss at Light Illusion

Author wikihua
ZRO

#18 | Posted: 14 Nov 2024 14:57 
Steve
Steve:
However, we are going to add Spectral Graphs to ColourSpace, so at that point we will natively use spectral data too.

That's great news about ColourSpace natively using spectral data!

Perhaps the original intention of this thread was to say that since XYZ/Yxy values are caculated by multiplying the measured spectral power distribution by the CMF function and then integrating. Having direct access to the spectral power distribution would allow for the flexibility of applying different CMFs (such as the CIE 1964 10° CMF) to convert into XYZ/Yxy values accordingly .

Author liyi_1991 Male
ZRO

#19 | Posted: 15 Nov 2024 21:29 
There will be no difference between calculating XYZ by colourspace and calculating XYZ by the spectrometer internal program. My guess is that calculating XYZ will slow down the operation of ColourSpace or increase instability.
Not all spectrometers provide CMF switching function, and software calculation can make up for this.
As a leading color science software, I think CS should not be limited to XYZ but trace back to the source of XYZ and calculate everything.
TOPCON SR-UL2 ︱ InstrumentSystems CAS 140CT-151 ︱ UPRTEK MK550T ︱ PhotoResearch PR-670 ︱ ADMESY Hyperion

Author Steve Male
INF

#20 | Posted: 10 Dec 2025 16:42 
We've had a couple of questions on CMFs recently, probably due to Calman stating they now include CIE 2015 CMFs...

The issue is that CMFs can only be used with Spectral data when converting into XYZ data, as used by most calibration systems.
This has been discussed within the thread already, but it is worth reiterating that trying to apply CMFs to existing XYZ measurement data is not a valid workflow, and will generate inaccurate results.
When using Spectros, many have the ability to internally apply a given CMS to the measured spectral data BEFORE exporting standard XYZ data to the calibration program.
This can be seen with CR-250/300 Spectros for example.

With Colourimeters, the action of Probe Matching effectively burns into the probe measurements the CMF that was in operation when the Spectro was use to generate the Probe Matching file.
With ColourSpace this can uniquely be applied retrospectively, rather than being burnt-in to the Colourimeter measurements.

As and when we add Spectral Data capabilities to ColourSpace we will at that point look at adding User Selective CMFs, so they operate correctly.
We will not add fake CMF options to recorded XYZ data.

Steve
Steve Shaw
Mob Boss at Light Illusion

Author titleexaminer92 Male
ZRO

#21 | Posted: 28 Aug 2026 18:27 
Steve:
As and when we add Spectral Data capabilities to ColourSpace we will at that point look at adding User Selective CMFs, so they operate correctly.
Any plans to add Spectral Data capabilities to ColourSpace in the near future?

Author Steve Male
INF

#22 | Posted: 28 Aug 2026 20:36 
We have lots of plans, but don't comment on future developments are that not likely to be in the next release.

Steve
Steve Shaw
Mob Boss at Light Illusion

Author liyi_1991 Male
ZRO

#23 | Posted: 29 Aug 2026 02:02 
For individuals, visual matching is far superior to any universal CMF method (universal CMF includes CIE 1931, CIE 170, Apple CMF, etc.). For industrial production, universal CMF is the only option.
Until methods and instruments for measuring individual CMF become available, I do not trust any universal CMF. They are sometimes effective, and sometimes worse than 1931.
TOPCON SR-UL2 ︱ InstrumentSystems CAS 140CT-151 ︱ UPRTEK MK550T ︱ PhotoResearch PR-670 ︱ ADMESY Hyperion

Author Kuky
ZRO

#24 | Posted: 31 Aug 2026 16:47 
Steve:
That's irrelevant, as ColourSpace doesn't (presently) use spectral data.
Then why are you querying Admesy Hera for spectral data on every measurement?

Author Steve Male
INF

#25 | Posted: 31 Aug 2026 16:52 
The code is in there for all (most?) spectros, ready for when we do.

As stated, presently we do not use spectral data.
The only before would be for spectral graphs.

Steve
Steve Shaw